DSO150 signal line problem

Post Reply
tomtom38
Posts: 5
Joined: Sun Apr 08, 2018 7:53 pm

DSO150 signal line problem

Post by tomtom38 »

Hello,
first the device works, but since few days the signal linie is at the top of the display.
I do not calibrate c3 or c5.
VPOS reset and trigger have no effect.

https://ibb.co/m25cKx
RobNL
Posts: 122
Joined: Mon Dec 25, 2017 3:11 pm
Location: The Netherlands
Contact:

Re: DSO150 signal line problem

Post by RobNL »

Set your switch on DC, it's now on AC.
CU, Rob
jye1
Posts: 1221
Joined: Wed Feb 20, 2013 4:58 pm

Re: DSO150 signal line problem

Post by jye1 »

@ tomtom38:

Please adjust sensitivity lower (the number larger) and also move the left border triangle lower. 0.1V/DIV is too high for 3.3V signal.
tomtom38
Posts: 5
Joined: Sun Apr 08, 2018 7:53 pm

Re: DSO150 signal line problem

Post by tomtom38 »

All settings have no effect.
Normally with the V Pos setting the signal line have to be in the middle.
I think now that is a hardware damage
Once It got 11 Volt for few seconds .
What can die first with an overvoltage?!
tomtom38
Posts: 5
Joined: Sun Apr 08, 2018 7:53 pm

Re: DSO150 signal line problem

Post by tomtom38 »

The Problem is there is no negative Voltage.It seems U5 is damage
AlexS16
Posts: 6
Joined: Wed Apr 04, 2018 7:46 pm

Re: DSO150 signal line problem

Post by AlexS16 »

tomtom38 wrote:The Problem is there is no negative Voltage.It seems U5 is damage
Are you sure about that? my scope has the very same problem
jye1
Posts: 1221
Joined: Wed Feb 20, 2013 4:58 pm

Re: DSO150 signal line problem

Post by jye1 »

U5 could be die if over voltage was applied. Please check V- and AV- with multimeter. If it is gone then that's the problem.
tomtom38
Posts: 5
Joined: Sun Apr 08, 2018 7:53 pm

Re: DSO150 signal line problem

Post by tomtom38 »

very sure UB is on Pin 8 /U5
but Output is a low positiv Voltage.
@jye1 from witch manufacturer is the ICL
I bought ICL7660SCBA Super Voltage Converter Intersil SO-8 1pcs
I hope that is similar.
jye1
Posts: 1221
Joined: Wed Feb 20, 2013 4:58 pm

Re: DSO150 signal line problem

Post by jye1 »

ICL7660SCBA can replace ICL7660 and is better.
Post Reply