Input is massively attenuated

Post Reply
tazziedave
Posts: 2
Joined: Mon Sep 18, 2017 10:44 pm

Input is massively attenuated

Post by tazziedave »

Hi

Having assembled the Shell I have the following Voltages:

Code: Select all

Vsup   9.01V
V+     8.17V
Av+    5.11V
V-    -7.61V
Av-   -5.04V
V1     2.9mV
V2    -0.8mV
V3     5.3mV
V4     1.69V
The test signal (shown on another oscilloscope) is as follows
0.1V Test signal on other scope
0.1V Test signal on other scope
3.3V Test signal on other scope
3.3V Test signal on other scope
on the Shell it looks like:
0.1V Test signal on Shell
0.1V Test signal on Shell
3.3V Test signal on Shell
3.3V Test signal on Shell
No amount of adjustment on C3 and C5 made any difference.

Any suggestions?

Thanks

Dave
jan
Posts: 42
Joined: Sun Jun 25, 2017 8:38 am
Location: the netherlands

Re: Input is massively attenuated

Post by jan »

Dave ,

see my post "C5 calibration on new version"

use the settings from the manual , c3 can affect both settings ( 0.1 and 3.3 ) c5 has very low influence.
jye1
Posts: 1221
Joined: Wed Feb 20, 2013 4:58 pm

Re: Input is massively attenuated

Post by jye1 »

@Dave:

In the photos the couple switch was set to GND. Please change it to DC and check again.
tazziedave
Posts: 2
Joined: Mon Sep 18, 2017 10:44 pm

Re: Input is massively attenuated

Post by tazziedave »

:oops: Yep, was the couple switch - obvious really. Just hadn't changed it after doing the Vpos alignment.

Maybe the manual could have a quick reminder in the adjusting section.

Thanks for your help.

Dave