Test is not square

Chev
Posts: 10
Joined: Wed Jun 17, 2015 1:21 pm

Test is not square

Post by Chev »

Hello everybody.

First, sorry for my english.

I finally made it work, reviewing the contact on LCD.
But when I do the manual test, the signal I get is not square, is the image.

I connected a 10 volt source (AC) and gives me the signal of the image.

Moreover, the response time is very slow, since I put a battery, for example, until the signal is displayed.

Can you help me?

Thank you.

Chev.
Attachments
DSC02528 si.JPG
DSC02526 si.JPG
davef
Posts: 78
Joined: Sat Jun 06, 2015 10:01 am

Re: Test is not square

Post by davef »

Can you display the built-in test signal?
Chev
Posts: 10
Joined: Wed Jun 17, 2015 1:21 pm

Re: Test is not square

Post by Chev »

davef wrote:Can you display the built-in test signal?
I'm sorry, I do not understand you, Davef. :oops:

The test signal (square) does not appear on the screen.

It appears what you see in the picture.

Thank you.
davef
Posts: 78
Joined: Sat Jun 06, 2015 10:01 am

Re: Test is not square

Post by davef »

Sorry, I didn't realise that you were trying to look at the built-in test signal!

Well, either the test signal is not on the test output or you have some issues in your input circuit.

I was also confused by the top waveform. If you expecting 50Hz and getting that waveform what timebase setting are you on? Bit hard to see but it looks like 0.2 seconds. Try 10 or 20ms.

Also your 9V battery will struggle to supply enough current for any length of time. What is the voltage at the DC input when the 'scope is runnng?
Chev
Posts: 10
Joined: Wed Jun 17, 2015 1:21 pm

Re: Test is not square

Post by Chev »

Thank you Davef.
davef wrote: Well, either the test signal is not on the test output or you have some issues in your input circuit.
And how do I fix it?
davef wrote: I was also confused by the top waveform. If you expecting 50Hz and getting that waveform what timebase setting are you on? Bit hard to see but it looks like 0.2 seconds. Try 10 or 20ms.
I've tried various combinations, 10 and 20 ms, but does not solve anything.
davef wrote: Also your 9V battery will struggle to supply enough current for any length of time. What is the voltage at the DC input when the 'scope is runnng?
I'll watch the battery current and I tell you.
The voltage, around 8,5, more or less.

Thank you again.
davef
Posts: 78
Joined: Sat Jun 06, 2015 10:01 am

Re: Test is not square

Post by davef »

Test signal ... if it is not on TESTSIG end of R26 you have microprocessor problems. You will need another oscilloscope to look for it.

How about measuring all the other voltages around the board.

When you had the 10VAC source on it what amplitude do you read on the screen, even though you can't see a good sinewave?
Chev
Posts: 10
Joined: Wed Jun 17, 2015 1:21 pm

Re: Test is not square

Post by Chev »

davef wrote:Test signal ... if it is not on TESTSIG end of R26 you have microprocessor problems. You will need another oscilloscope to look for it.
:? :roll: :cry:
davef wrote:How about measuring all the other voltages around the board.
All are close to the manual, but I would have to check them again to be sure.
davef wrote:When had the 10VAC source on it what amplitude do you read on the screen, even though you can't see a good sinewave?
I can not check it now, but I will and I'll tell you.

Thanks Davef.
davef
Posts: 78
Joined: Sat Jun 06, 2015 10:01 am

Re: Test is not square

Post by davef »

:? :roll: :cry:

Well, there can only be an open track from pin 17 (TESTSIG) on the uP to the end of R26 or a short on this track. Do a close visual inspection to confirm pin 17 is soldered to the PCB track and that there are no shorts on the track to R26. R26 could be open and/or you have shorted pin 1 of J2 to ground. Or your test probe could be faulty.

If it was my unit I would do a continuity test on the track and check for shorts to ground and that R26 was good. But, it is not my unit! Hope you understand where I am coming from.
Chev
Posts: 10
Joined: Wed Jun 17, 2015 1:21 pm

Re: Test is not square

Post by Chev »

Thank you Dave.

I will breathe and I'll get on it. ;)
Chev
Posts: 10
Joined: Wed Jun 17, 2015 1:21 pm

Re: Test is not square

Post by Chev »

davef wrote::? :roll: :cry:

Well, there can only be an open track from pin 17 (TESTSIG) on the uP to the end of R26 or a short on this track. Do a close visual inspection to confirm pin 17 is soldered to the PCB track and that there are no shorts on the track to R26. R26 could be open and/or you have shorted pin 1 of J2 to ground. Or your test probe could be faulty.

If it was my unit I would do a continuity test on the track and check for shorts to ground and that R26 was good. But, it is not my unit! Hope you understand where I am coming from.
Dave.

These are the data I obtained:
all voltages are almost exactly like manual, except D2, on the positive, instead of 9.39 mark, giving 8.71.

About battery:
8.92 volt.
3 mA

There are continuity between pin17 and R26 (right where it is written R26).

What else I can check? :oops: :)