Thank you for your reply.
MNTech wrote:V2 is the last test point before the signal enters the microprocessor. The voltage at V2 with no input should be approximately 1.6V. My unit measures 1.7V.
Cool, I'm getting 1.66V at V2
MNTech wrote:Set the switches to DC, 1V, X5.
Touch the red test lead to end of D2 (battery voltage) near the edge of the board.
The voltage at V1 should be 1/100th of the input with the switches set as above. The 9V battery drops with load so expect about 0.08V at that point.
Hmm, with the switches set as you describe touching the red lead to D2 seems to make no difference. I am getting 0V at "V1" and 1.66V at "V2" regardless of whether I touch the red lead to D2 or not.
I then checked the continuity of the lead and was surprised to discover the red end was not connected to the central connection of the BNC plug at all. The center of the BNC socket in the board is connected to the wire coming out the back but the center of the plug on the lead is not connected to the red alligator clip. The black alligator clip
is connected to the outside of the BNC socket (and ultimately all the way to "DGND") but the red is not connected to the center. I assume it should be?
Anyway, if I directly connect a lead from the wire coming out the back of the BNC socket to D2 I do get 92mV at "V1" and 1.8V at "V2" (0.14 higher than before). This certainly makes it look like I have a faulty lead. As it is I had actually already ordered a proper oscilloscope lead (with an attenuator) so will try that when it arrives.
Thank you for your help,
(: