Re: The bottom of the wave is being deformed

I assume you use a 10:1 probe?
Please read the topic about "Measurement errors with x10 probe on large negative signals",
why it happens and how it can be solved.
by Arnoldus
Wed Aug 08, 2018 6:02 pm
Forum: DSO Shell
Topic: [Firmware?] Sine wave reading is different in AC and DC mode
Replies: 0
Views: 3693

Re: [Firmware?] Sine wave reading is different in AC and DC

I assume you use a 10:1 probe?
It looks like the same problem as described in the following topic:
"Measurement errors with x10 probe on large negative signals".
by Arnoldus
Tue May 15, 2018 2:36 am
Forum: DSO Shell
Topic: Measurement errors with x10 probe on large negative signals
Replies: 13
Views: 21907

Re: Measurement errors with x10 probe on large negative sign

Hallo Mat,

Ik zal de voorgestelde testen doen, al heb ik zelf al diverse testen gedaan en kon ik geen afwijkingen constateren.
Maar ik ben op dit moment alleen ~7500 Km van huis....

Wordt volgende maand vervolgt,
Groeten,
Arnold
by Arnoldus
Fri Feb 23, 2018 3:38 pm
Forum: DSO Shell
Topic: DSO 15001K, LED D1 flashes after boot before assembly
Replies: 2
Views: 4444

Re: DSO 15001K, LED D1 flashes after boot before assembly

I had exactly the same (running FW110 and HW ver E ) but assembled it and works fine :)
by Arnoldus
Wed Feb 21, 2018 9:08 pm
Forum: DSO Shell
Topic: Noise and voltage reading problem
Replies: 6
Views: 7652

Re: Noise and voltage reading problem

May I give you a tip? I use this '9V' 300mAh NiMH battery (8,4V) from Conrad without any problem: https://www.conrad.nl/nl/oplaadbare-9v-batterij-blok-ansmann-maxe-6lr61-nimh-84-v-300-mah-1-stuks-1305467.html It remains most of the discharge time above 8V, and it runs at least 2 hours before chargin...
by Arnoldus
Wed Feb 21, 2018 8:44 pm
Forum: DSO Shell
Topic: Problem with calibrating
Replies: 1
Views: 3414

Re: Problem with calibrating

Hello Sidey, And the resistance from ground (batt-) to pin 2&5? Must be 0 ohm. It looks like the ground pin of the Rotary PCB does not make very well contact, becouse you say it works when you press on it. The track of both PCB's has a large copper area, so you have to solder it long enough to m...
by Arnoldus
Wed Jan 17, 2018 6:50 pm
Forum: DSO Shell
Topic: Negative offset
Replies: 1
Views: 3666

Re: Negative offset

It looks like you have the same problem as I described in post "DC offset at AC coupling with >3,5V signal".
I wrote a solution in the post about "Measurement errors with x10 probe on large negative signals",
because it is coused by the same input circuit problem.
by Arnoldus
Thu Jan 11, 2018 9:08 pm
Forum: DSO Shell
Topic: Where is the trigger point?
Replies: 1
Views: 3371

Re: Where is the trigger point?

Just "Hold down [SEC/DIV] button for about 3 seconds. This will put the triggerpoint at the center"
by Arnoldus
Wed Jan 10, 2018 10:01 pm
Forum: DSO Shell
Topic: DC offset at AC coupling with >3,5V signal
Replies: 2
Views: 3681

Re: DC offset at AC coupling with >3,5V signal SOLVED!

Problem fixed! See description on the last post concerning "measuring large negative signals with an 1:10 probe". in short: I used an OMRON G3VM61B1 solid state relais instead of an LDR. It has a FET and is only 2 ohm when ON, the LED needs only 3mA and the switch is bidirectional! I put t...
by Arnoldus
Wed Jan 10, 2018 9:53 pm
Forum: DSO Shell
Topic: Measurement errors with x10 probe on large negative signals
Replies: 13
Views: 21907

Re: Measurement errors with x10 probe on large negative sign

Hello, I just tried an other (easier) way to fix the problem, and used an OMRON G3VM61B1 solid state relais instead of an LDR. It has a FET and is only 2 ohm when ON, the LED needs only 3mA and the switch is bidirectional! I put the output side (FET) in series with the input of the opamp pin 12, jus...